Gas sensor loop current or voltage at the sensor head is set as per the manufacturer’s recommended value.
-
Gas monitor zero is adjusted to the instrument air.
-
A test gas with a known quantity of combustion gas (Generally methane 2% by volume) is fed to the sensor and the span is adjusted to read 40% on the monitor scale (refer to the station drawings for the exact settings parameters).
-
The calibration procedure is repeated until the zero and span reads correctly.
1 Like