How Often Should an Ammonia Gas Detector Be Bump Tested?

At minimum, the ammonia gas detector should be bump tested once a month as part of the monthly maintenance plan for the device. There is no limit on the number of bump tests; they may be done more often depending on application and the comfortability/confidence level one has in the device and how it responds. If a bump test fails, a full calibration should be done. Bump test dates and results should be written down in a log book.